112
4.2.3.5. Initiating a GPT Calibration Gas Generation
Note
It is highly recommended to perform a GPT Pre-Set before initiating any GPT gas
generation (Section 4.2.4).
To initiate GPT gas generation you will need to know:
The TOTAL GAS FLOW for the mixture output;
The Target O
3
concentration (equal to the target NO
2
concentration to be
generated), and;
The NO source gas concentration.
Then, press:
EXIT discards the new gas type &
target concentration
ENTR accepts the new gas type &
target concentration
STANDBY GPT:0.0 PPB O3
0 0 0 .0 PPB ENTR EXIT
Make sure that the
T700 is in STANDBY
mode
STANDBY SYSTEM RESET
AUTO MAN PURG GPT GPTPS
STANDBY A-CAL=0.0000 LPM
<TST TST> GEN STBY SEQ SETUP
GPT ACT CAL=2.000 LPM
TEST GEN STBY SEQ MSG CLR SETUP
STANDBY GPT:0.0 PPB NO
0 0 0 .0 PPB ENTR EXIT
Toggle these buttons
to set the NO target
concentration.
MUST be at least 10%
Higher than the Target
O
3
Concentration
STANDBY TOTAL FLOW = 2.000 LPM
02.000 ENTREXIT
Toggle these buttons to
set the target TOTAL
FLOW.
(Default = 2.000 LPM)
EXIT discards the new
flow rate
ENTR accepts the
new gas flow rate
The T700 will stay in generate mode
until the STBY button is pressed.
EXIT discards the new gas type &
target concentration
ENTR accepts the new gas type &
target concentration
Toggle these buttons
to set the NO target
concentration.
Should be equal to the
expected NO
2
concentration