Teledyne T700 Microphone User Manual


 
Getting Started Teledyne API – Model T700 Dynamic Dilution Calibrator
90
3.4.8. SELECTING AN OPERATING MODE FOR THE O3 GENERATOR
The O
3
generator can be set to operate in three different modes: Constant, Reference,
and Bench.
3.4.8.1. CNST (CONSTANT)
In this mode, the O
3
output of the generator is based on a single, constant, drive voltage.
There is no Feedback loop control by the T700’s CPU in this mode.
3.4.8.2. REF (REFERENCE)
The O
3
control loop will use the generator reference detector's UV lamp measurement as
input. This mode does not use the photometer to control the ozone generator.
This setting will be the default mode (if not equipped with the photometer option) of the
T700 calibrator and will be used whenever the calibrator is using the GENERATE
AUTO command or the GENERATE sequence step to create a calibration mixture.
When the GENERATE MAN command or the MANUAL sequence steps are active,
the local O
3
generator mode (chosen during when the command/step is programmed)
will take precedence.
3.4.8.3. BNCH (BENCH)
The O
3
concentration control loop will use the photometer’s O
3
measurement as input.
To select a default O
3
generator mode, press:
06873B DCN6388