T700, M700E Calibrator Manuals APPENDIX A-2: Setup Variables For Serial I/O, Revision D.3 (05623D DCN5839)
A-15
Setup Variable Numeric
Units
Default
Value
Value
Range
Description
REF_DELAY Seconds 60 1–300
O
3
generator reference feedback
control delay.
REF_FREQ Seconds 1 1–60
O
3
generator reference
adjustment frequency.
REF_FSIZE Samples 4 1–10 O
3
generator reference filter size.
REF_INTEG — 0.1 0–10
O
3
generator reference PID
integral coefficient.
REF_DERIV — 0.2 0–10
O
3
generator reference PID
derivative coefficient.
BENCH_DELAY Seconds 120 1–300
O
3
generator bench feedback
control delay.
BENCH_FREQ Seconds 10 1–60
O
3
generator bench adjustment
frequency.
BENCH_FSIZE Samples 3 1–10 O
3
generator bench filter size.
BENCH_INTEG — 0.2 0–10
O
3
generator bench PID integral
coefficient.
BENCH_DERIV — 0.5 0–10
O
3
generator bench PID
derivative coefficient.
DRIVE_STABIL mV 10 0.1–100
O
3
generator drive stability limit
to update concentration cache.
CACHE_RESOL PPB 2 0.1–20
O
3
generator cache un-
normalized concentration
resolution.
O3_LAMP_CYCLE Seconds 2 0.5–30
O
3
generator lamp temperature
control cycle period.
O3_LAMP_PROP 1/DegC 0.2 0–10
O
3
generator lamp temperature
PID proportional coefficient.
O3_LAMP_INTEG Gain 0.01 0–10
O
3
generator lamp temperature
PID integral coefficient.
O3_LAMP_DERIV Gain 0.2 0–10
O
3
generator lamp temperature
PID derivative coefficient.
25 MFC_PRESS_LIMIT PSIG
Warnings:
15–36
0–50
MFC pressure warning limits. Set
point not important.
20
8
5
REG_PRESS_LIM PSIG
Warnings:
15–25
5-11
5
0–50
Regulator pressure warning
limits. Set point not important.
50 PERM_SET1 ºC
Warnings:
49–51
0–100
Permeation tube #1 temperature
set point and warning limits.
50 PERM_SET2
2
ºC
Warnings:
49–51
0–100
Permeation tube #2 temperature
set point and warning limits.
TARGET_FLOW Lpm 2 0.01–20.00
Default total output flow rate, if
flow not specified in individual
actions or steps.