Teledyne T700 Microphone User Manual


 
Principles of Operation Teledyne API – Model T700 Dynamic Dilution Calibrator
290
10.3.7. POWER SUPPLY AND CIRCUIT BREAKER
The T700 calibrator operates in two main AC power ranges: 100-120 VAC and 220-240
VAC (both ± 10%) between 47 and 63 Hz. A 5-ampere circuit breaker is built into the
ON/OFF switch. In case of a wiring fault or incorrect supply power, the circuit breaker
will automatically turn off the calibrator.
WARNING
The T700 calibrator is equipped with a universal power supply that allows it to accept
any AC power configuration, within the limits specified in Table 2-2.
Should the power circuit breaker trip correct the condition causing this situation before
turning the calibrator back on.
SENSOR SUITES
LOGIC DEVICES
(e.g. CPU, I
2
C bus,
Motherboard, etc.)
RELAY
PCA
ON / OFF
SWITCH
PS 2
(+12 VDC)
OPTIONAL VALVES
GPT valve,
O
3
Gen valve
Photometer M/R
valve, etc.)
Cooling
Fan
PS 1
ANALOG SENSORS
O
3
Generator
Reference detector,
Photometer UV
Detector
Pre-Amplifiers
& Amplifiers
Sensor Control
& I/O Logic
Solenoid
Drivers
KEY
AC POWER
DC POWER
AC
POWER IN
+5 VDC
±15 VDC
O
3
Generator UV
Lamp Xformer
Photometer
UV Lamp P/S
GAS
PRESSURE
SENSORS
GAS
TEMPERATURE
SENSORS
VALVE
DRIVE PCA
DILUENT
VALVE
PURGE
VALVE
CAL GAS 1
VALVE
CAL GAS 2
VALVE
CAL GAS 3
VALVE
CAL GAS 4
VALVE
Solenoid
Drivers
MFC1
(Diluent)
MFC2
Cal Gas
MFC3
2nd Cal Gas
(Optional)
O
3
Generator UV
Lamp P/S
O
3
Generator
UV Lamp
Photometer
Pump
Figure 10-9: T700 Power Distribution Block diagram
06873B DCN6388