Teledyne API – Model T700 Dynamic Dilution Calibrator Calibration and Verification
209
7.3.6. O3 PHOTOMETER GAS FLOW CALIBRATION
Note
A separate flow meter is required for the procedure.
To calibrate the flow of gas through the T700 calibrator’s optional photometer bench.
1. Turn OFF the T700 Dynamic Dilution Calibrator.
2. Attach the flow meter directly to the EXHAUST port of the T700 calibrator.
3. Turn the T700 Dynamic Dilution Calibrator ON.
4. Perform the following steps: