Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator
128
4.3.1.6. Setting Up Control Outputs for a Sequence
The T700 calibrator’s control outputs allow the calibrator to control devices that accept
logic-level digital inputs, such as programmable logic controllers (PLCs), data loggers,
or digital relays/valve drivers.
They can be used as:
12 separate ON/OFF switches assigned to separate calibration sequences, or;
A 12-bit wide bus allowing the user to define activation codes for up to 4095
separate calibration sequences.
They can be set to:
Be active whenever a particular calibration sequence is operating, or;
Activate/deactivate as individual steps within a calibration sequence are run (see
Section 4.3.2.8).
To assign a CC OUTPUT pattern/code to a particular sequence, press: