Teledyne T700 Microphone User Manual


 
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation
123
4.3.1.3. Setting the Repeat Count for a Sequence
The sequence can be set to repeat a certain number of times, from 1 to 100. It can also
be set to repeat indefinitely by inputting a zero (0) into the REPEAT COUNTER.
To set the REPEAT COUNTER, press:
SETUP X.X REPEAT COUNT:1
<SET SET> EDIT EXIT
SETUP X.X REPEAT COUNT:[0]
00 1 ENTEREXIT
EXIT discards the
new REPEAT
COUNT
ENTR accepts the
new REPEAT
COUNT
Toggle these buttons to set the repeat count from 1 to 100.
Enter “0” to cause the sequence to loop indefinitely
Continue pressing SET> until ...
STANDBY A-CAL=0.000 LPM
<TST TST> GEN STBY SEQ SETUP
Make sure that the T700 is
in standby mode.
SETUP X.X PRIMARY SETUP MENU
GAS SEQ CFG CLK PASS MORE EXIT
SETUP X.X SEQUENCE CONFIGURATION
EDIT PRINT EXIT
SETUP X.X END OF SEQUENCES
INS PRNT EXIT
SETUP X.X 1) SEQ [NAME], [X] STEPS
PREV NEXT INS DEL EDIT PRNT EXIT
SETUP X.X NAME:0
SET> EDIT EXIT
Deletes the sequence shown
in the message field
Edits the sequence shown
in the message field
Scrolls back and forth between
existing sequences
This display only appears if there are no sequences currently
programmed into the T700.
OTHERWISE ...
06873B DCN6388