Communications Setup and Operation Teledyne API – Model T700 Dynamic Dilution Calibrator
188
5. In the instrument’s SETUP>MORE>COMM>COM2 menu, make the following settings:
Baud Rate: 115200
COM2 Mode Settings:
Quiet Mode ON Computer Mode ON
MODBUS RTU OFF MODBUS ASCII OFF
E,8,1 MODE OFF E,7,1 MODE OFF
RS-485 MODE OFF SECURITY MODE OFF
MULTIDROP MODE OFF ENABLE MODEM OFF
ERROR CHECKING ON XON/XOFF HANDSHAKE OFF
HARDWARE HANDSHAKE OFF HARDWARE FIFO ON
COMMAND PROMPT OFF
6. Next, configure your communications software, such as APIcom. Use the COM port
determined in Step 4 and the baud rate set in Step 5. The figures below show how
these parameters would be configured in the Instrument Properties window in
APIcom when configuring a new instrument. See the APIcom manual (PN 07463)
for more details.
Note
USB configuration requires that the baud rates of the instrument and
the PC match; check the PC baud rate and change if needed.
Using the USB port disallows use of the rear panel COM2 port except
for multidrop communication.