Teledyne T700 Microphone User Manual


 
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation
169
4.11.5.1. Creating a GENERATE LEVEL
To create a LEVEL using the T700’s AUTO generation function, press:
SETUP X.X GENERATE:0.0 PPB CO2
0.000PCTCO2ENTR EXIT
SETUP X.X GENERATE:ZERO
ZERO ENTR EXIT
Toggle this button to
scroll through the
available gas types (as
programmed during
initial setup.
Continue pressing this key until the desired
gas type appears
SETUP X.X GENERATE:0.0 PPB CO2
0 0 0 .0 PPB CO2 ENTR EXIT
Toggle this button
to to scroll through
the available units
of measure
SETUP X.X TOTAL FLOW = 2.000 LPM
0 2. 0 0 0 ENTR EXIT
EXIT discards the
new flow rate
Toggle these
buttons to set the
target
concentration.
Toggle these buttons to
set the target TOTAL
FLOW.
(Default = 2.000 LPM)
ENTR accepts the new gas flow rate
STANDBY LEVEL:0
00 ENTREXIT
Toggle these buttons
until the designation of
the existing defined level
program is reached.
EXIT discards the new
LEVEL number
ENTR accepts the new
LEVEL number
CHOOSE ACTION Submenu
SETUP X.X ACTION TO PERFORM:GENERATE
PREV NEXT ENTR EXIT
Starting at the CHOOSE ACTION Submenu
06873B DCN6388