4.11.5.4. Creating a MANUAL LEVEL
To create a level using the T700’s MANUAL generation function, press:
SETUP X.X INSERT STEP: MANUAL
ENTR EXIT
CHOOSE ACTION Submenu
SETUP X.X ACTION TO PERFORM:GENERATE
PREV NEXT ENTR EXIT
Use the NEXT until ...
Starting at the CHOOSE ACTION Submenu
SETUP X.X DILUENT GAS FLOW: 0.000 LPM
0 6 .0 0 0 ENTR EXIT
SETUP X.X CAL GAS TYPE:ZERO
ZERO ENTR EXIT
Toggle this button to
scroll through the
available gas types (as
programmed during
initial setup.
Continue pressing this key until the desired
gas type appears
SETUP X.X CAL GAS FLOW: 0.000 LPM
3 .0 0 0 0 ENTR EXIT
Toggle these
buttons to set the
target
concentration.
Toggle these buttons
to set the target
GAS
FLOW
.
EXIT discards the new
flow rate
ENTR accepts the
new gas flow rate
SETUP X.X CAL GAS TYPE:SO2
SO2 ENTR EXIT
SETUP X.X O3 GEN SET POINT: 000.0 PPB
0 0 0 0 .0 ENTR EXIT
SETUP X.X O3 GEN MODE: OFF
OFF CNST REF BNCH ENTR EXIT
This button
turns the O
3
Generator
OFF/ON
REF: The concentration control
loop will use the generator’s
reference detector as input.
BNCH: The concentration
control loop will use the
photometer bench.
SETUP X.X O3 GEN SET POINT: 0.0 MV
0 0 0 0 .0 ENTR EXIT
This button sets a CONSTANT drive
voltage for the O
3
Generator
These buttons set a target
concentration for the O
3
Generator
Toggle these
buttons to set the
CONSTANT drive
voltage of the O
3
generator
Toggle these keys to
set output
CONCENTRATION
of the O
3
generator
EXIT discards the new
setting
ENTR accepts the
new setting
STANDBY LEVEL:0
0 0 ENTR EXIT
Toggle these buttons
until the designation of
the existing defined level
program is reached.
EXIT discards the new
LEVEL number
ENTR accepts the new
LEVEL number