Verifying O3 Generator Performance
Using the set up shown in Figure 8-4, perform the following steps:
STANDBY GENERATE:0.0 PPB O3
0 400PPBO3ENTREXIT
Make sure that the
T700 is in STANDBY
mode
STANDBY SYSTEM RESET
AUTO MAN PURG
STANDBY GENERATE:ZERO
ZERO ENTR SETUP
STANDBY A-CAL=0.0000 LPM
<TST TST> GEN STBY SEQ SETUP
GENERATE ACT CAL=2.000 LPM
<SET SET> GEN STBY SEQ SETUP
Toggle this button to
scroll through the
available gas types (as
programmed during
initial setup.
Continue pressing this key until the desired
gas type appears
STANDBY GENERATE:0.0 PPB O3
0 000PPBO3ENTREXIT
Toggle these buttons
to set the target
concentration.
STANDBY TOTAL FLOW = 2.000 LPM
0 2. 0 0 0 ENTR EXIT
Toggle these buttons to
set the target TOTAL
FLOW.
(Default = 2.000 LPM)
Toggle this button
to set the units of
measure.
Record O
3
concentration from reference
photometer
Note
The readings recorded from the T700’s A-CAL test function and the reference
photometer should be within 1% of each other.