4.3.2.3. The GPTPS Step
This step operates and is programmed similarly to the GENERATE GPTPS (see
Section 4.2.4 for information on choosing the correct input values for this step).
At the end of
the programming sequence, the T700 firmware will automatically insert a
DURATION step that needs to be defined.
To insert a GPTPS step into a sequence, press:
EXIT discards the new gas type &
target concentration
ENTR accepts the new gas type &
target concentration
SETUP X.X GPTPS:0.0 PPB O3
0 0 0 .0 PPB ENTR EXIT
SETUP X.X GPTPS:0.0 PPB NO
0 0 0 .0 PPB ENTR EXIT
Toggle these buttons
to set the NO target
concentration.
MUST be at least 10%
Higher than the Target
O
3
Concentration
SETUP X.X TOTAL FLOW = 2.000 LPM
0 2. 0 0 0 ENTR EXIT
Toggle these buttons to
set the target TOTAL
FLOW.
(Default = 2.000 LPM)
EXIT discards the new
flow rate
ENTR accepts the
new gas flow rate
EXIT discards the new gas type &
target concentration
ENTR accepts the new gas type &
target concentration
Toggle these buttons
to set the NO target
concentration.
Should be equal to the
expected NO
2
concentration
INSERT STEP Submenu
SETUP X.X INSERT STEP: GENERATE
PREV NEXT ENTR EXIT
Starting at the STEPS Submenu
Use the PREV and NEXT keys to scroll though the
list of available instructions
SETUP X.X INSERT STEP: GPTPS
PREV NEXT ENTR EXIT
SETUP X.X 3) DURATION : 10.0 MIN
PREV NEXT INS DEL EDIT EXIT
SETUP X.X DURATION: 10.0 MIN
0 1 0 .0 ENTR EXIT
Toggle these
buttons to set
DURATION of this
step
EXIT discards the new
setting
ENTR accepts the
new setting