Calibration and Verification Teledyne API – Model T700 Dynamic Dilution Calibrator
220
7.5.1.1. Calibrating the Diluent, Cal Gas Optional O
3
Generator Pressure Sensors
1. Turn off the calibrator and open the top cover.
2. For the sensor being calibrated, insert a “T” pneumatic connector at the location
described in Table 7-2 and shown in Figure 7-6 and Figure 7-7.
3. Turn o
n the calibrator and perform the following steps:
4. Turn OFF the T700.
5. Remove the pressure monitor.
6. Restore the pneumatic lines to their proper connections.
7. Close the calibrator’s cover.
06873B DCN6388