Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator
164
To adjust the signal levels of an analog output channel manually, press:
From the
AIO CONFIGURATION SUBMENU
(See Section 6.9.1.1.)
DIAG ANALOG I/O CONFIGURATION
PREV NEXT ENTR EXIT
Continue pressing SET> until you reach the
output to be configured
DIAG AIO TEST_OUTPUT: 5V, CONC2, NOCAL
<SET SET> EDIT EXIT
Continue pressing SET> until ...
DIAG AIO TEST_OUTPUT: CALIBRATED:NO
<SET SET> CAL EXIT
DIAG AIO TEST_OUTPUT: RANGE: 5V
SET> EDIT EXIT
DIAG AIO TEST_OUTPUT: VOLT-Z: 0 mV
U100 UP10 UP DOWN DN10 D100 ENTREXIT
DIAG AIO TEST_OUTPUT: VOLT-S: 4500 mV
U100 UP10 UP DOWN DN10 D100 ENTREXIT
DIAG AIO TEST_OUTPUT: CALIBRATED: YES
<SET SET> CAL EXIT
These menus
only appear if
AUTO-CAL is
turned OFF
These buttons increase /
decrease the analog output
signal level (not the value on the
display)
by 100, 10 or 1 counts.
Continue adjustments until the
voltage measured at the output
of the analyzer and/or the input
of the recording device reads 0
mV or 90% of full scale.
DIAG AIO AOUTS CALIBRATED: NO
SET> CAL EXIT
06873B DCN6388