60
CONNECTING THE CALIBRATOR TO A SAMPLE GAS MANIFOLD
Use this setup when connecting the T700 calibrator to an analyzer network using a
sample manifold. In this case, the sampling cane and the manifold itself act as the vent
for the T700.
&
gas outlets
be capped
T700 output flow rate
must be greater than the
requirements
of the entire system,
whichever is higher.
Minimum O.D. of this gas line
must be 3/8 inch
Figure 3-18: Set up for T700 – Connecting the T700 to a Sample Manifold
Note
• This is the recommended method for connecting the T700 calibrator to a
system with analyzers that DO NOT have internal zero/span valves.
• The manifolds as shown in the above drawing are oriented to simplify the
drawing. Their actual orientation in your set-up is with the ports facing
upward. All unused ports must be capped.
• When initiating calibration, wait a minimum of 15 minutes for the calibrator to
flood the entire sampling system with calibration gas.