Overview of Operating Modes and Basic Operation Teledyne API – Model T700 Dynamic Dilution Calibrator
170
4.11.5.2. Creating a GPT LEVEL
To create a LEVEL using the T700’s GPT function, press:
Use the NEXT until ...
EXIT discards the new gas type &
target concentration
ENTR accepts the new gas type &
target concentration
GPT GPT:0.0 PPB O3
0 0 0 .0 PPB ENTR EXIT
SETUP X.X INSERT STEP: GPT
PREV NEXT ENTR EXIT
GPT GPT:0.0 PPB NO
0 0 0 .0 PPB ENTR EXIT
Toggle these buttons
to set the NO target
concentration.
MUST be at least 10%
higher than the target O
3
Concentration
GPT TOTAL FLOW = 2.000 LPM
0 2. 0 0 0 ENTR EXIT
Toggle these s to set the
target TOTAL FLOW.
(Default = 2.000 LPM)
EXIT discards the new
flow rate
ENTR accepts the
new gas flow rate
EXIT discards the new gas type &
target concentration
ENTR accepts the new gas type &
target concentration
Toggle these buttons
to set the NO target
concentration.
Should be equal to the
expected NO
2
concentration
CHOOSE ACTION Submenu
SETUP X.X ACTION TO PERFORM:GENERATE
PREV NEXT ENTR EXIT
Starting at the CHOOSE ACTION Submenu
STANDBY LEVEL:0
0 0 ENTR EXIT
Toggle these buttons
until the designation of
the existing defined level
program is reached.
EXIT discards the new
LEVEL number
ENTR accepts the new
LEVEL number
06873B DCN6388