3.3.3. PERMEATION TUBE SETUP FOR THE T700
1. Press SETUP and GAS
2. Press PERM
3. Enter the elution rate for the permeation tube and select the type of gas by pressing the
gas button to scroll through the gas list until the desired gas is shown.
Note
The name of the gas produced by the permeation tube generator MUST be
different from the name of any bottle connected to the calibrator.
4. Then enter the gas flow through the permeation tube. This should be done with the flow
standard connected at the outlet of the perm tube oven.