Teledyne T700 Microphone User Manual


 
Communications Setup and Operation Teledyne API – Model T700 Dynamic Dilution Calibrator
182
Table 5-3: LAN/Internet Configuration Properties
PROPERTY
DEFAULT
STATE
DESCRIPTION
DHCP STATUS
On
This displays whether the DHCP is turned ON or OFF.
INSTRUMENT
IP ADDRESS
0.0.0.0
This string of four packets of 1 to 3 numbers each (e.g. 192.168.76.55.)
is the address of the calibrator itself.
GATEWAY IP
ADDRESS
0.0.0.0
A string of numbers very similar to the Instrument IP address (e.g.
192.168.76.1.) that is the address of the computer used by your LAN to
access the Internet.
SUBNET MASK
0.0.0.0
Also, a string of four packets of 1 to 3 numbers each (e.g.
255.255.252.0) that defines that identifies the LAN to which the device is
connected.
All addressable devices and computers on a LAN must have the same
subnet mask. Any transmissions sent devices with different subnet
masks are assumed to be outside of the LAN and are routed through a
different gateway computer onto the Internet.
TCP PORT
1
3000
This number defines the terminal control port by which the instrument is
addressed by terminal emulation software, such as Internet or Teledyne
API’s APICOM.
HOST NAME T700
The name by which your calibrator will appear when addressed from
other computers on the LAN or via the Internet. While the default setting
for all Teledyne API’s T700 calibrators is “T700”, the host name may be
changed to fit customer needs.
1
Do not change the setting for this property unless instructed to by Teledyne API’s Customer Service
personnel.
Note
If the gateway IP, instrument IP and the subnet mask are all zeroes (e.g.
“0.0.0.0”), the DCHP was not successful in which case you may have to
configure the calibrator’s Ethernet properties manually. See your network
administrator.
06873B DCN6388