Teledyne T700 Microphone User Manual


 
Teledyne API – Model T700 Dynamic Dilution Calibrator Getting Started
63
SPAN GAS/ZERO AIR CALIBRATION MANIFOLD EXHAUST/VENT LINES
The span and zero air manifolds’ excess gas should be vented to a suitable vent outside
of the room. This vent should be of large enough internal diameter to avoid any
appreciable pressure drop, and it must be located sufficiently downstream of the output
ports to assure that no ambient air enters the manifold due to eddy currents or back
diffusion.
Figure 3-21: T700 Pneumatic Diagram – Base Unit
The standard T700 Dynamic Dilution Calibrator is equipped with one calibration gas
mass flow controller (flow rate
0 – 100 cm
3
/min) and one diluent gas mass flow controller
(flow rate 0-10 LPM). See Table 1-1 for the various flow rate options.
3.3.2.6. Other Pneumatic Connections
Some of the T700 Dynamic Dilution Calibrator’s optional equipment requires additional
pneumatic connections.
O
3
GENERATOR OPTION
Because ozone (O
3
) quickly breaks down into molecular oxygen (O
2
), this calibration
gas cannot be supplied in precisely calibrated bottles like other gases such as SO
2
, CO,
CO
2
NO, H
2
S, etc. The optional O
3
generator extends the capabilities of the T700
Dynamic Dilution Calibrator dynamically generate calibration gas mixtures containing
O
3
.
Additionally a glass mixture volume, designed to meet US EPA guidelines for Gas
Phase Titration (GPT), is included with this option. This chamber, in combination with
06873B DCN6388