Calibration and Verification Teledyne API – Model T700 Dynamic Dilution Calibrator
216
7.4.2.4. Turning O
3
Generator Calibration Points ON / OFF
To enable or disable an existing calibration point, press:
STANDBY A-CAL=0.000 LPM
<TST TST> GEN STBY SEQ SETUP
Make sure that the T700 is
in standby mode.
SETUP X.X PRIMARY SETUP MENU
GAS SEQ CFG CLK PASS MORE EXIT
SETUP X.X SECONDARY SETUP MENU
COMM FLOW VA DIAG EXIT
SETUP X.X ENTER PASSWORD
818 ENTREXIT
Toggle these buttons to
enter the correct
PASSWORD
DIAG SIGNAL I/O
PREV NEXT ENTR EXIT
DIAG O3 GEN CALIBRATION
PREV NEXT ENTR EXIT
Continue pressing NEXT until ...
DIAG O3 GEN CALIBRATION
CAL PNTS EXIT
DIAG O3GEN 1) 500 MV, 5.0 MIN, ON
PREV NEXT INS DEL EDIT PRNT EXIT
Continue pressing PREV & NEXT until your
reach the point to be turned ON/OFF
DIAG O3GEN 8) 1500 MV, 5.0 MIN, ON
PREV NEXT INS DEL EDIT PRNT EXIT
DIAG O3GEN CAL. POINT DRIVE:0 MV
<SET SET> EDIT EXIT
DIAG O3GEN CAL. POINT ENABLELD:ON
<SET SET> EDIT EXIT
Continue pressing SET> until ...
DIAG O3GEN CAL. POINT ENABLELD:ON
ON ENTR EXIT
Toggle this button to turn
the point ON / OFF
EXIT discards
the new setting
ENTR accepts
the new setting
06873B DCN6388